关于热电偶的转接有专门的CTL决议CTL-OP_108_Ed.2.0.pdf,你的办法是被否决的,$ M7 R) |: C* S3 t3 O' V1 I
特别注意我高亮的一句话:& n: g9 `1 G/ y& n( H
3.8 Exception - Thermocouple connectors need not be used in the limited situation where the! n d7 H. O2 {8 S" ~9 |8 y8 G
thermoelectric affects of the junctions made are nullified such as shown in the following R7 X$ \( _* ^. f2 P
example for feed through connectors. Thermocouple wires TC#1 and TC#2 are the same type.
8 y. J T# ^ H8 {0 [5 hConnections are made to the same type of conductors (e.g. copper alloy). The temperatures
' Z8 b. N( `9 ]T1 and T2 at the connections are the same.: ^" i8 Y5 D& d9 M! ]% ?
CTL决议请参考:
, z9 g: I2 e- z. D4 i/ R/ e全文下载:http://www.eychi.com/upload/stan ... L-OP_108_Ed.2.0.pdf
! R- g8 a( M# C$ P8 ^- E* R/ T# W
$ [* A+ I3 z7 w1 v: v A决议截图:
+ W: n, i: j8 C9 B. v9 I$ A% D4 x; `/ N7 W
: R7 J4 w, d* U8 s/ t8 |& t
7 Z5 B- p2 I( N/ }如果你的做法可行,我们做仪器的就可以节约很多接线端子了,比如这个错误的做法:# _( a8 A9 S6 }& C' D- `8 G" ^5 K
. i9 O" |, R5 c8 W: b- A为节约热电偶接线端子,竟然使用转换开关,必须要9个接线端子,一个就忽悠了客户,这是坑客户的做法,1米测试角实际应用:16元/个×81个×3面=3888元,坑人做法:16元/个×18个=288元,节省3600元-10位开关费用,至少节约:2800元。3 }7 k+ Y2 ]: y8 k! H# S2 H* \0 [
: k X1 S9 f) E东莞安规实际生产1米×1米×1米的测试角会使用至少不少于243个接线端子(不含顶部板):7 x: @0 U0 T8 G$ I8 Y, Q
9 P8 K2 T3 U5 t' A0 O! P
可以看到,我们生产的时候是密密麻麻全是热电偶接线端子的。' n- v4 W+ J1 u- q# T+ V
再来一个巨无霸半成品测试角,也是密密麻麻布满了热电偶接线端子,总共479个接线端子,如果也像那样节省成本,我们赚大发了: