关于热电偶的转接有专门的CTL决议CTL-OP_108_Ed.2.0.pdf,你的办法是被否决的,0 ^% Y Z& k& ^ S3 b1 ~
特别注意我高亮的一句话:
Q1 T, h" O$ K: L ^3.8 Exception - Thermocouple connectors need not be used in the limited situation where the
+ M( d8 g4 }( A; fthermoelectric affects of the junctions made are nullified such as shown in the following) B0 \. L* y8 k, W
example for feed through connectors. Thermocouple wires TC#1 and TC#2 are the same type.5 T, T6 K% n! W0 v1 D; |& o
Connections are made to the same type of conductors (e.g. copper alloy). The temperatures* Y6 ^* _/ {, m/ E
T1 and T2 at the connections are the same.! ]- B# D+ Z) m+ t$ @, o
CTL决议请参考:% X- p2 C z9 T# `; x
全文下载:http://www.eychi.com/upload/stan ... L-OP_108_Ed.2.0.pdf) G' Z' c; b c7 T
! ]2 a4 k4 w- O2 t# s7 ]6 {0 x
决议截图:
r5 D3 p; [9 b5 b' S K6 C8 M0 W+ g% v2 ]9 _
7 z6 C) f6 L o. L" @5 x9 R6 b
# b& W4 c" P4 L; A( F0 C; O如果你的做法可行,我们做仪器的就可以节约很多接线端子了,比如这个错误的做法:
" ?, a% s% t0 l* H7 ^0 b& ?$ H# X6 k
为节约热电偶接线端子,竟然使用转换开关,必须要9个接线端子,一个就忽悠了客户,这是坑客户的做法,1米测试角实际应用:16元/个×81个×3面=3888元,坑人做法:16元/个×18个=288元,节省3600元-10位开关费用,至少节约:2800元。& n+ q" c3 L* S
1 L ` ?3 z- p' C+ A$ D
东莞安规实际生产1米×1米×1米的测试角会使用至少不少于243个接线端子(不含顶部板):
7 |$ _# ~: v& d7 i 8 ?" g( k( P% B1 S& }3 b
可以看到,我们生产的时候是密密麻麻全是热电偶接线端子的。
3 ^) S5 O3 m% E r. Y6 o: R) b再来一个巨无霸半成品测试角,也是密密麻麻布满了热电偶接线端子,总共479个接线端子,如果也像那样节省成本,我们赚大发了:/ d( P/ l. l* H* i" j; x1 ?
$ H- j; c) q: S; u
+ t# c" u$ x0 q$ t U |