灯管用第二种方法测试电压24.2V,算出来有48.4mA。而我在实验室用228 表测试只有2.5mA. 不知原因出在哪里?现在负责的工程师坚持第二种方法的测试结果,因为标准根本没提到可以用228表测试。
+ h. n* V- [) v/ Y/ g. H/ B: t9 K: A( M1 V- p3 @# ]2 @- h1 L
* j$ V7 ]7 _& i3 f; g$ k
第一种方法:Readings were taken directly in MIU using a suitable leakage current meter (i.e.: Simpson 228 set to "let go", etc.).
9 h+ U# A# |, C9 Q
, e- G- a5 ]$ I$ C第二种方法:Readings were taken in millivolts using a suitable voltmeter connected across the test circuit defined in UL935, Figure 24.2. These readings were then converted to MIU by dividing them by 500.- E( m. k3 V5 l$ [8 m4 X A
|
|