安规网

 找回密码
 注册安规
安规论坛 | 仪器设备 | 求职招聘 | 国家标准 公告 | 教程 | 家电 | 灯具 | 环保 | ITAV 签到 充值 在线 打卡 设备 好友| 帖子| 空间| 日志| 相册
IP淋雨机 | 证书查询 | 规范下载 | 资质查询 招聘 | 考试 | 线缆 | 玩具 | 标准 | 综 合 红包 邮箱 打卡 工资 禁言 分享| 记录| 道具| 勋章| 任务
水平垂直燃烧机 | 针焰 | 灼热丝 | 漏电起痕
IP防水防尘设备|拉力机|恒温恒湿|标准试验指
灯头量规|插头量规|静风烤箱|电池设备|球压
万年历 | 距中秋节还有
自2007年5月10日,安规网已运行
IP淋雨设备| 恒温恒湿箱| 拉力机| 医疗检测设备沙特Saber 埃及COI 中东GCC|CoC直接发证机构水平垂直燃烧机|灼热丝|针焰试验机|漏电起痕试验机
灯头量规|试验指|插头插座量规|灯具检测设备耐划痕试验机|可程式恒温恒湿试验箱 | 耦合器设备广东安规-原厂生产-满足标准-审核无忧
12
返回列表 发新帖
楼主: zdream_80
打印 上一主题 下一主题

[综合话题] 测试电压选取的由来?(60601-1,9706.1)

[复制链接]
11#
发表于 2009-3-6 12:52 | 只看该作者
广东安规检测
有限公司提供:
Equipment is operated at rated voltage 请注意看这个话中的rated voltage.是额定电压下测试,而不是正常状态。! x4 W5 o  G1 i6 q. _, J3 ]
另外可以参考下UL的test datasheet 中的描述9 b9 W, I' k1 \
The equipment was operated for 1 min at rated voltage or at the upper limit of the rated voltage.  Then the sample was de-energized by removal of the power cord plug from the supply source.  One second after disconnection the voltage between points specified below was measured.  A storage oscilloscope with probes having an impedance of at least 100 MΩ or similar high impedance instrument was used for this measurement.  The test was repeated ten times for each pair of the measurement points.
12#
发表于 2009-3-11 16:28 | 只看该作者
看看10.2.2就明白了.
13#
发表于 2009-3-12 09:51 | 只看该作者
FYI, only applicable to the part of capacitor discharge test:
" v5 k4 ^: h  wTo ensure patients and equipment operators are not exposed to hazardous energy levels, voltage levels are measured on all mains terminals accessible after disconnection. With the unit
' u$ H9 i, [9 ~+ ?operating at 110 % of the rated voltage, power is disconnected and the voltage decay at the mains connection is recorded. The voltage should decay to a level of 60 VDC or less within 1 second. The test is repeated ten times for each pair of the measurement points. The theory is that in those ten times, you will catch the AC waveform discharging at or close to the peak of the mains voltage cycle. It is at the peak where you will have the highest voltage and the longest decay
# w( c3 q- m* xtimes. An alternative method of establishing the discharge characteristics of the unit is simply just to monitor the waveform on an oscilloscope and keep connecting and disconnecting the power until a peak reading occurs. By this method, you can see visually when the AC power waveform has been shut off at the peak level and a single measurement can be recorded and reported.
  K) v$ Y4 p* V9 `- O6 f: @+ l, @) i" o* p6 U% v; ]8 v0 W+ x  {
Author: Steven Henderson is a Product Safety Manager at Curtis-Straus LLC.
14#
发表于 2009-3-12 15:59 | 只看该作者
请注意比较以下的两条来自标准的条款:
) I0 h% T/ h5 u9 h1 c- U. }3 k. c剩余电压:& N* h3 n+ }$ e
EQUIPMENT is operated at RATED voltage or at the upper limit of the RATED voltage range.
7 f5 L0 y/ \- J: @7 z7 |0 l$ ~漏电流测试:
2 U/ I8 G1 C9 m4 j0 [With a supply equal to 110% of the highest RATED MAINS VOLTAGE.
4 u( Z% ]: A! R6 N" Z同是rated voltage, 如何理解呢?
15#
发表于 2009-3-22 23:03 | 只看该作者
就IEC6061中来说,110%额定电压并不作为单一故障状态,而10.2.2中也说明了电压可波动10%,所以个人认为,本标准中所有与电压相关的实验都可用110%额定电压进行实验。
16#
发表于 2009-3-31 10:41 | 只看该作者
一般的国家电网正常波动在7%左右,标准规定10%可以满足正常波动的需要,只要电压的变化会造成试验结果的变化的项目,都必须考虑,例如漏电流、剩余电压、功率等等都应该考虑。
17#
发表于 2009-3-31 18:53 | 只看该作者
电网的波动按照10%来考虑,按照标准要求除了剩余电压,其他的漏电流、功率、超温等等都应该考虑+-10%,即:最高额定网电源电压的1.1倍或是0.9倍。。。
; \: g5 c' K( I; d! k; e) z& W: S1 l2 F. A; M
关于剩余电压标准中只要求是在额定电压下或是额定电压范围上限的电压运行。。。。
您需要登录后才可以回帖 登录 | 注册安规

本版积分规则

关闭

安规网为您推荐上一条 /1 下一条

QQ|关于安规|小黑屋|安规QQ群|Archiver|手机版|安规网 ( 粤ICP13023453-10 )

GMT+8, 2024-6-19 22:34 , Processed in 0.131401 second(s), 14 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.|广东安规赞助

快速回复 返回顶部 返回列表