应该是不用。
8 F/ N9 A) z, H ~, [part1 clause25.1 中提到这个测试是"between the terminal so one winding"这个应该是只要测 ...
; W# h4 y* M7 ?6 Y- Q" s! }6 ^+ e8 y( B, R$ N
这个测试项目的频率为400HZ,这是5085-2里的,而5085-3里的这项测试的要求是 the dielectric voltage withstand test after rated output heating test, twice the rated voltage was applied to the transformer's winding with the remaining [winding] open circuited. The frequency of the applied voltage potential was not less than twice the rated frequency and was maintained for a period of 7200 c. s, G6 S! X6 O; j' m
p% L. b/ z* E: D5 Y' b问题来了----为什么二者同一个测试项目区别这么大呢?7200C的含义是什么? |
|