Subject: Leakage current in secondary circuits
/ Q v( |5 W+ z5 N) hKey words: Leakage current, secondary circuit: U# B. l) u" m) ~1 Y6 p
Confirmed by CTL at its 39th meeting, in Cologne4 l4 H1 e& _. C1 ^5 A
Question:
- l* k& s4 |' oIf secondary circuit impedances limit the leakage current, is further investigation of secondary
* _; W J4 E g4 s; W; p" lcircuits required? (refer to sub-clause 52.5).
7 \$ D* g, D4 WDecision:
4 m* |( k% m# I8 }Secondary circuits providing protective means after short-circuiting of inadequate AIR
$ s: e0 ~. C7 P1 T( V* jCLEARANCE and CREEPAGE DISTANCE must be investigated. Failure of components in
, v. }3 O n5 Kthese circuits shall be investigated as a SINGLE FAULT CONDITION. Failure of such7 y# a/ J! j7 [
components shall be investigated as a SINGLE FAULT CONDITION.