灯管用第二种方法测试电压24.2V,算出来有48.4mA。而我在实验室用228 表测试只有2.5mA. 不知原因出在哪里?现在负责的工程师坚持第二种方法的测试结果,因为标准根本没提到可以用228表测试。5 r8 v9 G( P5 I7 G
$ d/ Z$ i) p: _$ _- l. }( J6 p( l+ e+ D& ~9 v
第一种方法:Readings were taken directly in MIU using a suitable leakage current meter (i.e.: Simpson 228 set to "let go", etc.).
/ u4 x4 v3 w7 Z$ K; y9 B' O* M! C+ f" q
第二种方法:Readings were taken in millivolts using a suitable voltmeter connected across the test circuit defined in UL935, Figure 24.2. These readings were then converted to MIU by dividing them by 500./ Q# i: B2 ]2 n7 j1 o
|
|