| Flammability rating of internal wiring
2 i# s Z+ O5 j7 n6 i# V0 c: ]2 z+ y | 9.2.1 a)
8 M. g _" i1 Y! ~ | 61010-1(ed.2)
( I' I9 o2 }- T: A0 [ |
, R- H! h+ p+ H& V6 d
# O. X9 Y' G8 J, h; g- CStandard(s)- (year and edition):
% i$ U, u; Q; ?. x% f: @" xIEC 61010-1, ed. 2, 2001: n- `8 `) S9 j n" y' z
Sub clause(s):* ?" D) l; F( s; ]2 H$ C5 L4 W
9.2.1 a)
$ p: D0 N7 H) ISheet n°: DSH 618
) Y7 [5 G' A2 H% E- z# D4 P7 c* wSubject:' u; Y9 n- \* J% ^
Flammability rating of internal wiring
" n. f# @1 m$ `4 V5 c' [, SKey words:
' D3 A/ d7 O$ R& ^+ o; [/ }Wires, flammability,5 m( u* ]. ^5 s2 |. T1 \
Decision approved by the CTL Plenary meeting 20076 n: c+ b: g/ e W6 X+ p0 S
Question:: H, W5 H+ W) `7 m* X# d
Insulated wires shall be equivalent FV-1 or better. Wires are not generally tested and rated according to FV-1. It is not easy to establish equivalence between different test classifications/methods if you are not involved directly in such testing procedures.
; X! L5 Y7 d( vWires typically are rated VW-1 or FT-1. It is unclear if these flammability ratings are equivalent and thus acceptable (Same problem exists with IEC 60601-1 ed. 3)3 a; e1 l1 I2 c: J6 S
Decision:! K7 P$ G8 ^ \6 _* z7 L9 x
Wires rated e.g. VW-1 or FT-1 or according to IEC 60332-1-2 are deemed equivalent and can be accepted.: p% y6 t- N* i y